Relationship Between Electrode-to-Modiolus Distance and Current Levels for Adults With Cochlear Implants.

Davis TJ, Zhang D, Gifford RH, Dawant BM, Labadie RF, Noble JH
Otol Neurotol. 2016 37 (1): 31-7

PMID: 26649603 · PMCID: PMC4675044 · DOI:10.1097/MAO.0000000000000896

HYPOTHESIS - Electrode-to-modiolus distance is correlated with clinically programmed stimulation levels.

BACKGROUND - Conventional wisdom has long supposed a significant relationship between cochlear implant (CI) stimulation levels and electrode-to-modiolus distance; however, to date, no such formal investigation has been completed. Thus, the purpose of this project was to investigate the relationship between stimulation levels and electrode-to-modiolus distance. A strong correlation between the two would suggest that stimulation levels might be used to estimate electrode-to-modiolus geometry.

METHODS - Electrode-to-modiolus distance was determined via CT imaging using validated CI position analysis software in 137 implanted ears from the three manufacturers holding FDA approval in the United States. Analysis included 2,365 total electrodes, with 1,472 from precurved arrays. Distances were compared to clinically programmed C/M levels that were converted to charge units.

RESULTS - Mean modiolar distance with perimodiolar and lateral wall electrodes was 0.47 and 1.15 mm, respectively. Mean suprathreshold charge values were significantly different between each manufacturer. When combining all data, we found a moderate positive correlation (r = 0.367, p < 0.01) that was driven both by the different charge values across companies, and that the company with the highest mean charge values only offers straight electrode arrays. When grouped by electrode type, however, we found a weak correlation (r = 0.12, p < 0.01) for perimodiolar array electrodes only. When considering a single array type from any one manufacturer, only one was observed where distance mildly predicted charge.

CONCLUSION - Our results suggest that electrode distance minimally contributes to the current level required for suprathreshold stimulation.