Multisite, multivendor validation of the accuracy and reproducibility of proton-density fat-fraction quantification at 1.5T and 3T using a fat-water phantom.

Hernando D, Sharma SD, Aliyari Ghasabeh M, Alvis BD, Arora SS, Hamilton G, Pan L, Shaffer JM, Sofue K, Szeverenyi NM, Welch EB, Yuan Q, Bashir MR, Kamel IR, Rice MJ, Sirlin CB, Yokoo T, Reeder SB
Magn Reson Med. 2017 77 (4): 1516-1524

PMID: 27080068 · PMCID: PMC4835219 · DOI:10.1002/mrm.26228

PURPOSE - To evaluate the accuracy and reproducibility of quantitative chemical shift-encoded (CSE) MRI to quantify proton-density fat-fraction (PDFF) in a fat-water phantom across sites, vendors, field strengths, and protocols.

METHODS - Six sites (Philips, Siemens, and GE Healthcare) participated in this study. A phantom containing multiple vials with various oil/water suspensions (PDFF:0%-100%) was built, shipped to each site, and scanned at 1.5T and 3T using two CSE protocols per field strength. Confounder-corrected PDFF maps were reconstructed using a common algorithm. To assess accuracy, PDFF bias and linear regression with the known PDFF were calculated. To assess reproducibility, measurements were compared across sites, vendors, field strengths, and protocols using analysis of covariance (ANCOVA), Bland-Altman analysis, and the intraclass correlation coefficient (ICC).

RESULTS - PDFF measurements revealed an overall absolute bias (across sites, field strengths, and protocols) of 0.22% (95% confidence interval, 0.07%-0.38%) and R  > 0.995 relative to the known PDFF at each site, field strength, and protocol, with a slope between 0.96 and 1.02 and an intercept between -0.56% and 1.13%. ANCOVA did not reveal effects of field strength (P = 0.36) or protocol (P = 0.19). There was a significant effect of vendor (F = 25.13, P = 1.07 × 10 ) with a bias of -0.37% (Philips) and -1.22% (Siemens) relative to GE Healthcare. The overall ICC was 0.999.

CONCLUSION - CSE-based fat quantification is accurate and reproducible across sites, vendors, field strengths, and protocols. Magn Reson Med 77:1516-1524, 2017. © 2016 International Society for Magnetic Resonance in Medicine.

© 2016 International Society for Magnetic Resonance in Medicine.